Topic 10 · Laboratory & Sampling
Set B Questions 11 to 20
- 11
Chain of custody documentation records
Explanation
It establishes that the sample was not tampered with, which matters when results are used legally.
- 12
Samples for most wastewater analyses must be preserved at
Explanation
Chilling slows the biological activity that would otherwise change the sample before it is analysed.
- 13
Holding time for a sample refers to the maximum period between
Explanation
Exceeding the holding time invalidates the result regardless of how carefully the test was run.
- 14
The pH of a sample must be measured
Explanation
Carbon dioxide exchange with the air begins changing the pH almost immediately after sampling.
- 15
A pH meter should be calibrated using
Explanation
Two or more buffers bracket the expected range and confirm that the electrode responds correctly.
- 16
The dissolved oxygen membrane probe should be calibrated
Explanation
Membrane fouling and electrolyte depletion both cause drift that a fresh calibration corrects.
- 17
The Winkler method is a laboratory procedure used to determine
Explanation
It is a titration method and is often used to verify the accuracy of a membrane probe.
- 18
Chemical oxygen demand differs from biochemical oxygen demand because it
Explanation
The result is available in hours rather than days, but it includes matter that organisms cannot use.
- 19
The ratio of BOD to COD in a domestic wastewater is normally about
Explanation
A ratio that changes suddenly often signals an industrial discharge of non biodegradable material.
- 20
Settleable solids are reported from an Imhoff cone reading as
Explanation
The cone is read after one hour of settling and the result is expressed as a volume.