Topic 07 · Water Chemistry & Lab
Set C Questions 21 to 30
- 21
What is a matrix spike used for in laboratory quality assurance and quality control?
Explanation
A matrix spike adds a known quantity of target analyte to an environmental sample. Calculating percent recovery reveals whether substances in the water interfere with or suppress the test method.
- 22
What is the term for the duplicate analysis of a single homogenized sample to assess precision?
Explanation
Duplicate sample testing analyzes two separate aliquots of the exact same sample. The Relative Percent Difference (RPD) between the two results measures the precision and repeatability of the method.
- 23
What primary chemical reagent is used in the amperometric titration of chlorine residuals?
Explanation
Amperometric titration uses phenylarsine oxide (PAO) as the reducing titrant. It measures electrical current changes at a noble metal electrode and is the most accurate method for free/combined chlorine.
- 24
What is the threshold odor number (TON) test used for in drinking water laboratory analysis?
Explanation
The TON test dilutes an odorous sample with odor-free water until a panel of testers can no longer detect odor. TON = (sample volume + odor-free volume) / sample volume.
- 25
What primary form of alkalinity exists in drinking water with a pH between 7.0 and 8.3?
Explanation
In the typical drinking water pH range of 7.0 to 8.3, nearly all alkalinity exists as bicarbonate (HCO3-). Carbonate appears above pH 8.3, while hydroxide is negligible below pH 10.0.
- 26
What is the secondary Maximum Contaminant Level guideline for iron and manganese in drinking water?
Explanation
EPA Secondary Drinking Water Standards set non-enforceable aesthetic limits of 0.3 mg/L for iron (causes red/brown staining) and 0.05 mg/L for manganese (causes black staining and bitter taste).
- 27
What analytical method is standard for measuring fluoride concentration in drinking water labs?
Explanation
Fluoride is routinely measured using an Ion Selective Electrode (ISE) with TISAB buffer (to eliminate iron/aluminum interferences) or by SPADNS colorimetric spectrophotometry.
- 28
What is the optimal fluoride concentration recommended by the US Public Health Service for dental health?
Explanation
The recommended target concentration for community water fluoridation is 0.7 mg/L. The enforceable EPA Primary MCL is 4.0 mg/L, and the Secondary SMCL is 2.0 mg/L (to prevent dental fluorosis).
- 29
What does a Total Organic Carbon (TOC) analyzer measure in water samples?
Explanation
TOC measures the carbon content of dissolved and particulate organic matter (humic acids, tannins, organics) by converting all organic carbon into CO2 via catalytic combustion or UV-persulfate oxidation.
- 30
What chemical reaction occurs during the standard Winkler titration method for dissolved oxygen?
Explanation
The Winkler method fixes DO with manganous sulfate and alkali-iodide-azide reagent to form a precipitate. Acidification releases free iodine proportional to DO, which is titrated with standard thiosulfate.